Current readings on new ANAN-7000's with serial numbers ending in "A"
Posted: Tue Aug 16, 2022 2:17 pm
Unless and until a software fix in Thetis can be implemented, anyone with a new ANAN-7000 with a serial number ending in "A" will not be able to obtain correct current readings in Thetis.
The reason for this is Apache chose to replace the ACS713-30A current sensor on the RF board with the TMCS1100A2 due to part obsolescence. This has the unfortunate effect of causing all current readings to be low by a factor of 0.55, i.e. 2 amps will read as 1.1