S Meter issue ( FIXED!)

User avatar
DL2XY
Posts: 116
Joined: Sun Jul 30, 2017 9:47 am

Re: S Meter issue ( not fixed )

Postby DL2XY » Sun Jan 08, 2023 9:01 pm

Or you may try to patch the database directly.
Setting the values to zero would reset the calibration.

Database RX Calibration.png
Database RX Calibration.png (69.18 KiB) Viewed 2677 times
User avatar
w-u-2-o
Posts: 5570
Joined: Fri Mar 10, 2017 1:47 pm

Re: S Meter issue ( not fixed )

Postby w-u-2-o » Sun Jan 08, 2023 10:32 pm

IIRC, there are many, many more places in the database than those that are touched by the calibration routine.
Chiron
Posts: 11
Joined: Sat Feb 26, 2022 8:00 am

Re: S Meter issue ( not fixed )

Postby Chiron » Tue Jan 10, 2023 10:31 pm

ramdor wrote:
Chiron wrote:Can it been solved by using (going back to) 2.9.0.5 and do the calibration and save it. Then go to version 2.9.0.6 again?


yah probably, but have not tried myself



Hello Richie,
I can confirm that using your latest master branch 2.9.0.6 and compile it, the generated 2.9.0.7 does store the calibrated value.
I can also see some Smeter additions. After succesfull storing the s-meter calibration, I can start version 2.9.0.6 again with the correct S-meter values.

My S-meter calibration problem is solved now. :D
User avatar
WM4CH
Posts: 59
Joined: Wed Nov 09, 2022 1:20 pm
Location: Cape Cod, MA

Re: S Meter issue ( not fixed )

Postby WM4CH » Sat Mar 11, 2023 2:48 pm

I thought I would update this.

I had some family health issues that took priority for a few months and I finally sent the radio to Doug last week and got it back yesterday.

The previous owner had overloaded the radio and the low pass filter on the RX1 path had blown so the Rx1 signal was down about 30db.

Doug replaced the filter and the protection diode which was not blown. So it appears the previous owner overloaded the Rx1 blew out the protection diode and the low pass filter, replaced the protection diode which didn't fix it and then sold me the radio. This is conjecture, but the only plausible situation where the low pass filter would be blown but the protection diode was not. The diode had obviously been replaced before by the looks of it. The bottom line the radio was sold to me damaged.

Doug fixed and I am now on the air with it. Doug does great work with a very fast turn around time.
User avatar
w-u-2-o
Posts: 5570
Joined: Fri Mar 10, 2017 1:47 pm

Re: S Meter issue ( FIXED!)

Postby w-u-2-o » Sat Mar 11, 2023 3:34 pm

Great news! :)

It's always shocking how many dishonest hams there are. You would think our hobby would attract a better class of citizen, but that's been proven far from true many times.

Doug is the man!
User avatar
WM4CH
Posts: 59
Joined: Wed Nov 09, 2022 1:20 pm
Location: Cape Cod, MA

Re: S Meter issue ( FIXED!)

Postby WM4CH » Sat Mar 11, 2023 4:34 pm

Yeah, Doug is great to work with
User avatar
n2gq
Posts: 139
Joined: Sun Dec 18, 2022 7:26 pm

Re: S Meter issue ( not fixed )

Postby n2gq » Wed Apr 05, 2023 2:40 pm

DL2XY wrote:
N3GX wrote:. By default Thetis reports about 7dBm in excess, (a calibrated -73 turns into -66 dBm.)


Do not calibrate at -73dB !
This is an ADC-System, ADC's need to be calibrated near fullscale. -20dBm should be fine.

Calibrating at S9 means 1/1000 of fullscale voltage, so the software must extrapolate by a factor of 1000 meaning all measurement errors are multiplied by this factor too.


Hi
Would -43 ok for a lev calibration?

Jay n2gq
Jay N2GQ
User avatar
DL2XY
Posts: 116
Joined: Sun Jul 30, 2017 9:47 am

Re: S Meter issue ( not fixed )

Postby DL2XY » Wed Apr 05, 2023 7:57 pm

n2gq wrote:Hi
Would -43 ok for a lev calibration?

Jay n2gq


-43dBm (S9+30) should be sufficient.
-20dBm(S9+53) would be still better since at least15bits of the 16bit ADC are used.

Walter dl2xy
User avatar
n2gq
Posts: 139
Joined: Sun Dec 18, 2022 7:26 pm

Re: S Meter issue ( not fixed )

Postby n2gq » Wed Apr 05, 2023 8:38 pm

DL2XY wrote:
n2gq wrote:Hi
Would -43 ok for a lev calibration?

Jay n2gq


-43dBm (S9+30) should be sufficient.
-20dBm(S9+53) would be still better since at least15bits of the 16bit ADC are used.

Walter dl2xy


Hi Walter
Okay I was going to purchase a s9 signal generator which has 3 different outputs. -43 being the highest, -73 s9 in the middle and -104 for s4 signal. They are for sale on ebay. Was also waiting on a TinySa ultra but they are still out of stock.

Thank you
Jay n2gq
Jay N2GQ
User avatar
n2gq
Posts: 139
Joined: Sun Dec 18, 2022 7:26 pm

Re: S Meter issue ( FIXED!)

Postby n2gq » Fri May 05, 2023 3:32 pm

When calibrating the s-meter is a 1k tone used or just unmodulated carrier?
Jay N2GQ
User avatar
w-u-2-o
Posts: 5570
Joined: Fri Mar 10, 2017 1:47 pm

Re: S Meter issue ( FIXED!)

Postby w-u-2-o » Fri May 05, 2023 4:13 pm

Just a carrier.

A tone would never be necessary for any sort of RF power measurement

In SSB there would be no difference between a 1KHz tone and a carrier except for the difference between the center frequency of the tone and dial frequency.

Just make sure the carrier is reasonably centered in the passband.
User avatar
n2gq
Posts: 139
Joined: Sun Dec 18, 2022 7:26 pm

Re: S Meter issue ( FIXED!)

Postby n2gq » Fri May 05, 2023 4:26 pm

Okay Scott
Thank you....
Jay N2GQ
User avatar
KA9UVY
Posts: 41
Joined: Thu Mar 02, 2023 5:58 am

Re: S Meter issue ( FIXED!)

Postby KA9UVY » Sat May 06, 2023 6:27 am

Hey Jay, Be sure and let us know if the S9 signal generator works out for you... I tried an old signal generator I had here and it failed to work. I see the S9 sig generator has some different freq available but none of which are on even Mhz.
Thetis didn't let me put in anything other than the Mhz digit.?
Maybe I am missing the adjustment but no matter what I type in it jumps to the next even Mhz.

73, Robert
KA9UVY
Anan 7000DLE- MKIII
Thetis 2.10.3.5 u2
FW 2.2.2A
Protocol 2 v 3.9

73, Robert
KA9UVY
User avatar
n2gq
Posts: 139
Joined: Sun Dec 18, 2022 7:26 pm

Re: S Meter issue ( FIXED!)

Postby n2gq » Sat May 06, 2023 1:53 pm

KA9UVY wrote:Hey Jay, Be sure and let us know if the S9 signal generator works out for you... I tried an old signal generator I had here and it failed to work. I see the S9 sig generator has some different freq available but none of which are on even Mhz.
Thetis didn't let me put in anything other than the Mhz digit.?
Maybe I am missing the adjustment but no matter what I type in it jumps to the next even Mhz.

73, Robert
KA9UVY


Hi Robert,
I changed direction and purchased a TinySA Ultra which can output -19dbm and you can set it to whatever frequency we use. I set the TinySa to -20dbm and then changed the value in Thetis to -20dbm as well as frequency and then press start.
Jay N2GQ
User avatar
n2gq
Posts: 139
Joined: Sun Dec 18, 2022 7:26 pm

Re: S Meter issue ( FIXED!)

Postby n2gq » Sun May 07, 2023 12:07 am

KA9UVY wrote:Hey Jay, Be sure and let us know if the S9 signal generator works out for you... I tried an old signal generator I had here and it failed to work. I see the S9 sig generator has some different freq available but none of which are on even Mhz.
Thetis didn't let me put in anything other than the Mhz digit.?
Maybe I am missing the adjustment but no matter what I type in it jumps to the next even Mhz.

73, Robert
KA9UVY


If you decide to purchase, go to RandL electronics. They have the legitimate device.
Jay N2GQ
User avatar
KA9UVY
Posts: 41
Joined: Thu Mar 02, 2023 5:58 am

Re: S Meter issue ( FIXED!)

Postby KA9UVY » Fri May 12, 2023 4:40 am

Thanks Jay, I purchased a tiny SA and calibrated mine this evening.

The tiny SA will be handy to have here anyways, thanks for the tip on it.
73, Robert
KA9UVY
Anan 7000DLE- MKIII
Thetis 2.10.3.5 u2
FW 2.2.2A
Protocol 2 v 3.9

73, Robert
KA9UVY
rdwing
Posts: 94
Joined: Sat May 21, 2022 2:05 am

Re: S Meter issue ( FIXED!)

Postby rdwing » Fri May 12, 2023 10:45 pm

The TinySA Ultra is packed with lots of great features, including the SA, signal generator, zero span, waterfall/fft, tons of useful built in measurement tools like OIP3, phase noise, harmonics, thd, channel power, and more. I use it with appropriate taps and attenuators for taking calibrated power measurements too. Can't more highly reccomend getting one. R and L is the place.
User avatar
n2gq
Posts: 139
Joined: Sun Dec 18, 2022 7:26 pm

Re: S Meter issue ( FIXED!)

Postby n2gq » Sat May 13, 2023 12:09 am

KA9UVY wrote:Thanks Jay, I purchased a tiny SA and calibrated mine this evening.

The tiny SA will be handy to have here anyways, thanks for the tip on it.
73, Robert
KA9UVY



Cool, glad it worked for you as well.
Jay N2GQ